Test measurements on 10 channels of the multiplexer

Go down

Test measurements on 10 channels of the multiplexer

Post  Alessandro on Wed Oct 28, 2009 7:27 pm

For this test I considered the following 10 channels 1-2-3-4-5 and 17-18-19-20-21.
For each channel I put a load resistance.



In the next image we can see how you have set the Channels:



Then, using the method of the "scan" (Scan Count: 5, Measure Count: 1) I measured the V for the first 5 resistance
and R for the others (See column II(measure) and VII(channels)).
The results are shown in the table below and are saved in a file.



Measure of the value of resistance for all channels (See column II(measure) and VII(channels)):



Next test:
-
Measures with the use of thermometers
- Set the acquisition system to make repeated measurements with the method of the trigger at each time interval.
- Optimization of the save file data.
avatar
Alessandro

Posts : 42
Join date : 2008-12-11
Location : Roma

View user profile

Back to top Go down

Test scan every 10 sec (10 channels) for 2 minutes

Post  Alessandro on Fri Oct 30, 2009 12:13 am

Setting scan:
- Scan Count: 12
- Measure Count: 1

Setting Simple Trigger:



In this way, every 10s, will start the scan.
All 10 channels are read simultaneously, In this case we measure the value of the 10 resistors:



It could be read simultaneously also their voltage, but will require an additional 10 channels (so the channels will be 20).
In our case, each thermometer will occupy two channels!

Important Points:
-
Must set the Scan Count and Measure Count
-
Must set the Simple Trigger
- Create a save file data
- We can also decide the start time for the scan
- Finally, save the data